peak at 4.2 min, nobody can tell me what it is, ideas

Question Slow Clap ×1

Method: C18 150×4.6, 5µm. 0.1% TFA in water / acetonitrile, 20→45% over 25 min. 214nm, 20µL injection. Semaglutide, reseller vial, nothing exotic.

Main peak at 11.8 min as expected. There is a reproducible 0.9% peak at 4.2 min that I cannot place. It survives a blank subtraction, so it is not the mobile phase, and it is in every injection from this vial.

Ideas welcome. I have run out of them.

257 up / 75 down77% upvoted18 commentsid 2aahjf26 Jul 2026

18 comments

13 in this archive, depth 6

best — the order this archive was captured in

u/column_dead_volume267 points·3 days ago

4.2 min on that gradient is very early — you are basically still at 20% organic. That is where you find things that barely retain on C18: salts, small polar fragments, excipients, and occasionally the preservative if the vial was reconstituted.

Was it? Because benzyl alcohol elutes early and 0.9% of total area would not be surprising if you injected a reconstituted sample rather than redissolving powder in mobile phase.

replysharereportpermalink
u/gradient_goblinOP178 points·3 days ago

oh no. it was reconstituted. with bacteriostatic water.

replysharereportpermalink
u/column_dead_volume234 points·2 days ago

There it is. That is your benzyl alcohol. Redissolve dry powder in mobile phase and the peak will vanish.

replysharereportpermalink
u/gradient_goblinOP154 points·2 days ago·edited

ran it. gone. completely gone. i have spent four evenings on a preservative. edit: leaving this up in full because somebody else will do this and at least now they can find it.

replysharereportpermalink
u/sig_figs_samMOD132 points·1 days ago

Adding it to the c/hplc wiki under "early peaks that are not your product". This is the third time in two years and the first time anyone wrote it down.

replysharereportpermalink
u/blank_injection94 points·1 days ago

the blank did not catch it because the blank was mobile phase, not bac water. run a bac-water blank as well when the sample was reconstituted.

replysharereportpermalink
u/rt_shift_reggie121 points·2 days ago

my first guess was ghost peak from plumbing but a reproducible 0.9% that survives blank subtraction is too well behaved for that. plumbing ghosts wander.

replysharereportpermalink
u/mass_spec_maggieMS104 points·3 days ago

the general answer to "what is this peak" is always MS. UV gives you a retention time and a size, which narrows it to a family.

replysharereportpermalink
u/gradient_goblinOP66 points·3 days ago

i do not have MS access and that is the constraint on my entire hobby.

replysharereportpermalink
u/mass_spec_maggieMS88 points·3 days ago

send me a vial and i will run it on a lunch break. genuinely. modmail.

replysharereportpermalink
u/purity_pedantanalytical58 points·3 days ago

worth noting that if a vendor ran a reconstituted sample this way, their reported purity would be 0.9 points lower than the material deserves.

replysharereportpermalink
u/blank_injection47 points·3 days ago

run a blank. then run another blank. then post. i say this every week and it is finally load-bearing.

replysharereportpermalink
[removed]0 points·3 days ago

[removed by moderator]

replysharereportpermalink
Permalinked branches
Deep branches get their own page so a single reply chain can be linked and read on its own.
About c/hplc

The analytical deep end: RP-HPLC methods, gradient design, integration decisions, LC-MS confirmation, amino-acid analysis, water content, and why two labs can report 98.1% and 96.4% on the same vial and both be right. Post the chromatogram, not the summary line.

32kmembers
130submissions
Sep 2023created
submissions / month, last year
Sponsored

Janoshik Analytical

Independent HPLC and mass spec. The number you get is the number they found.

janoshik.com
Sponsored

GL Biochem (Shanghai)

Custom peptides and amino acids direct from the manufacturer since 1998. ISO 9001 / cGMP. Batch COA every order.

glbiochem.net
c/hplc rules
  1. Chromatogram or it did not happen. Axis labels included.
  2. State the method: column, gradient, detection wavelength, injection volume.
  3. Area% is not mass%. Posts that conflate them get a correction flair, not a removal.
  4. Independent community. Nobody here sells anything, and anyone who tries is banned.
  5. Not medical advice. Describe what you did; never prescribe to a stranger.
  6. Claims need evidence. Batch numbers, dated screenshots, independent test reports, or a citation.
  7. No referral links, discount codes or affiliate URLs. Permanent ban, no appeal.
  8. No contact handles, wallet addresses or tracking numbers — they identify people.
  9. Be recognisably decent. Disagree hard, insult nobody.
Moderators
Volunteers. Unpaid, unaffiliated, and reachable through modmail only.
Before you read on

Several compounds discussed on GLP Research Hub are sold for research use only and are not approved for human use anywhere. Nothing here is medical advice and none of it is written by your clinician. If a post reads like an instruction, treat it as a description of what one stranger did.