peak at 4.2 min, nobody can tell me what it is, ideas

Questionbranch of 6 comments

Method: C18 150×4.6, 5µm. 0.1% TFA in water / acetonitrile, 20→45% over 25 min. 214nm, 20µL injection. Semaglutide, reseller vial, nothing exotic. Main peak at 11.8 min as expected. There is a reproducible 0.9% peak at 4.2 min that I cannot place. It survives a blank subtraction, so it is not the mobile phase, and it…

Read the full submission and all 18 comments →

This branch

6 comments, started 3 days ago
u/column_dead_volume267 points·3 days ago

4.2 min on that gradient is very early — you are basically still at 20% organic. That is where you find things that barely retain on C18: salts, small polar fragments, excipients, and occasionally the preservative if the vial was reconstituted.

Was it? Because benzyl alcohol elutes early and 0.9% of total area would not be surprising if you injected a reconstituted sample rather than redissolving powder in mobile phase.

replysharereportpermalink
u/gradient_goblinOP178 points·3 days ago

oh no. it was reconstituted. with bacteriostatic water.

replysharereportpermalink
u/column_dead_volume234 points·2 days ago

There it is. That is your benzyl alcohol. Redissolve dry powder in mobile phase and the peak will vanish.

replysharereportpermalink
u/gradient_goblinOP154 points·2 days ago·edited

ran it. gone. completely gone. i have spent four evenings on a preservative. edit: leaving this up in full because somebody else will do this and at least now they can find it.

replysharereportpermalink
u/sig_figs_samMOD132 points·1 days ago

Adding it to the c/hplc wiki under "early peaks that are not your product". This is the third time in two years and the first time anyone wrote it down.

replysharereportpermalink
u/blank_injection94 points·1 days ago

the blank did not catch it because the blank was mobile phase, not bac water. run a bac-water blank as well when the sample was reconstituted.

replysharereportpermalink

← back to the whole thread

About c/hplc

The analytical deep end: RP-HPLC methods, gradient design, integration decisions, LC-MS confirmation, amino-acid analysis, water content, and why two labs can report 98.1% and 96.4% on the same vial and both be right. Post the chromatogram, not the summary line.

32kmembers
130submissions
Sep 2023created
submissions / month, last year
Sponsored

Sigma-Aldrich Standards

Certified reference materials for peptide identity and purity work.

sigmaaldrich.com
c/hplc rules
  1. Chromatogram or it did not happen. Axis labels included.
  2. State the method: column, gradient, detection wavelength, injection volume.
  3. Area% is not mass%. Posts that conflate them get a correction flair, not a removal.
  4. Independent community. Nobody here sells anything, and anyone who tries is banned.
  5. Not medical advice. Describe what you did; never prescribe to a stranger.
  6. Claims need evidence. Batch numbers, dated screenshots, independent test reports, or a citation.
  7. No referral links, discount codes or affiliate URLs. Permanent ban, no appeal.
  8. No contact handles, wallet addresses or tracking numbers — they identify people.
  9. Be recognisably decent. Disagree hard, insult nobody.
Moderators
Volunteers. Unpaid, unaffiliated, and reachable through modmail only.
Before you read on

Several compounds discussed on GLP Research Hub are sold for research use only and are not approved for human use anywhere. Nothing here is medical advice and none of it is written by your clinician. If a post reads like an instruction, treat it as a description of what one stranger did.